Inconsistencies in current waveforms published in : Levi et al. “Observation of abundant heat production from a reactor device and of isotopic changes in the fuel” (referred to as TPR2), and related : Levi et al. “Indication of anomalous heat energy production in a reactor device “(referred to as TPR1).
Andrea S. , 14 October 2014, [reedited 23 Nov 2014]
I hate to say this, but the plot in TPR2 figure 5 and page 6, although not on a readable scale, may be revealing.
Please refer to
http://www.cobraf.com/forum/immagini/R_123566844_1.pdf
for a better reading.
The PCE-830, when displaying voltage and current waveforms ,normally synchronizes all waveforms to the upwards zero-crossing of the V1 waveform when present, or of the I1 waveform if V1 is not available, and sets the timescale to one period.
The displayed I3 waveform is determined by the triacs closure. Current flows when the triac on V3 is closed at the same time as V1 or V2, and it is a fair assumption that all three triacs are commanded with the same conduction angle (since a single feedback is used for all).
Now the displayed waveform for I3, is not the one would expect, neither with clockwise nor with counter-clockwise three-wire three-phase connection
The plot in TPR2 figure 5, instead is well matched by a negative I3 plot in the CCW connection, such as would result from inverting the clamp ammeter. [Note: this statement has been challenged, as the I3 plot seems to trigger at the beginning of the timescale. I thank Franco (who writes on the Cobraf Forum) for his sound technical advice and S. Caggia (22passi blog) for his constructive criticism ].
Now, the effect of reverting the I3 probe would be to underestimate the actual input power by a factor 3, i.e. close to the declared COP.
[ A later analysis, admitting the downstream PCE-830 to be connected 4-wire by using the upstream neutral (although the neutral is not shown in the TPR2), does find a plausible reading that fits the peaks' orientation in the TPR2 figure 5. On the other hand, the same analysis shows that current pulsewidth in figure 5 is compatible with a 3KW consumption (implying COP around 1), whereas the 0.9kW consumption declared by the authors would yield much narrower pulses. The updated analysis can be found in
http://www.cobraf.com/forum/immagini/R_123571297_1.pdf .
Once more the authors should better describe their setup, clarify the instance when the plot was taken, clarify if the measurement was 4-wire, and justify the apparent 3KW consumption resulting. The reversed clamp hypothesis remains the one that best fits the three-fold power underestimation .]
It does appear absurd that such a gross mistake could have happened. What is most disturbing is that a similar discrepancy on the PCE-830 plots happened in the TPR1.
In the TPR1 the I1 waveform similarly appeared inverted with respect to the one expected, such as would result by inverting the current probe.
When simulating the circuit on a spreadsheet, with triac conduction angle set for low regimes (as in the experiment setup) the probe reversal results in overestimating the input power.
But when the triac conduction angle is set for high regimes (as in the experiment run), the probe reversal results in underestimating the input power by a factor 2.7, close to the claimed COP.
Conclusion: the discrepancies in the published current waveforms, unless otherwise explained, hint to a gross mistake in misplacing a clamp ammeter as a plausible explanation for the apparent energy gain in both TPR1 and TPR2. The authors, for their own sake and that of the scientific community, are invited to provide sufficient evidence (additional photos, raw data, non-ambiguous explanations) that such an occurrence can be excluded.