That's good to know. Thanks. I think that's what the instrument company told Mizuno, which is why he selected the hot wire one.
As I said, they recently ran a comparison with a vane type. The vane one results were a little higher. It is a little hard to compare them, but I think I am doing it right. It was maybe 16% higher, worst case. That would increase the excess heat. That can't be right, because a 16% increase would make some of the low-power calibrations show excess heat, which is impossible. They may have done some more tests with the new anemometer. I assume they did, but I don't have any additional data from it. I expect if they spent a week or two wringing it out, they could find out why it was a little higher, fix the problem, and reconcile the numbers. Maybe not. Maybe they would have to redesign the orifice? I wouldn't know. Of course you would have to re-calibrate and redo everything from the get-go if you did that.
Even if the two anemometers do not agree completely, I am encouraged. There would have to be gigantic error to "explain away" these results. They would have to be wrong by factor of 6. A 16% error cannot begin to explain that.
Jed, the 16% error is most likely due to the reduced area that the air has to flow around the center of the anemometer cowling. Since the air flow now has to flow through the annulus space between the cowling and outside piping, the air velocity can be slightly increased when measuring flow from small orifices like Mizuno's calorimeter effluent pipe. This is irrelevant when measuring larger HVAC diffusors/AHUs where these instruments are normally used. But when measuring flow from a small pipe, the path the air flow has to take can increase apparent velocity.